The scan area of the chip does not fit, what to do?

With the new ISAC 112 slides, a customer had problems getting a full picture of the 4 images, e.g. not all four squares were scanned properly (the grid does not cover the first line). What is the reason and what do I do?

As the new ImmunoCAP ISAC 112 has a new chip design as well you need to update the ROI (region of interest) for the scanning. To do this you need to select the following coordinates in MIA: Scan Images -> Settings -> IgE -> X = 1400 / Y = 2200 / width = 11000 / height = 37000