What does this error message mean and what can I do about it? “Calibration curve warning Different PMT on calibration and measurement slides. Do you want to proceed anyway?”

What does this error message mean and what can I do about it? “Calibration curve warning Different PMT on calibration and measurement slides. Do you want to proceed anyway?”
This message means that the latest calibration curve that was run had another PMT value than you have in your scanner at the moment. Probably the PMT value has been “trimmed in” by service personnel after your latest run of calibration curve. Your options are: 1) Run a new calibration curve and proceed with running your samples again (recommended) 2) Click “yes” in the message box and you will get your results. BUT this means that you may get results that are not optimal since they will be measured against the “old” calibration curve that was run with another PMT value.